I have a different interpretation of this case. I suspect you did not tighten the set screw all the way in order to allow sliding of the lag screw and fracture fragment compression, as you would for a standard intertrochanteric hip fracture. This is a reverse obliquity fracture, so the result is migration of the proximal segment. I doubt reaming the canal would have prevented this. Had the set screw been fully tightened, there would be no postoperative fragment motion.
|
|
|